CCP Oxygenator & Pump Maintenance 2 — Questions and Answers
Question 1: Which parameter is most commonly used to assess oxygenator gas transfer efficiency during cardiopulmonary bypass?
- Transmembrane pressure gradient
- Arterial oxygen saturation and PaO2 (Correct answer)
- Sweep gas flow rate alone
- Oxygenator blood prime volume
Correct answer: Arterial oxygen saturation and PaO2
Arterial oxygen saturation and PaO2 directly reflect the oxygenator's ability to transfer oxygen into the blood.
Question 2: During CPB, a sudden increase in oxygenator inlet-to-outlet pressure gradient with no change in flow rate most likely indicates:
- Venous line obstruction
- Oxygenator fiber bundle clotting or thrombosis (Correct answer)
- Air embolism in the arterial line
- Pump head failure
Correct answer: Oxygenator fiber bundle clotting or thrombosis
Progressive clotting within the hollow-fiber bundle increases resistance and raises the transmembrane pressure gradient.
Question 3: What is the primary function of the heat exchanger integrated within a modern membrane oxygenator?
- Removing CO2 from perfusate
- Regulating patient blood temperature during bypass (Correct answer)
- Filtering microemboli from arterial blood
- Increasing oxygen partial pressure
Correct answer: Regulating patient blood temperature during bypass
The heat exchanger uses countercurrent water flow to warm or cool blood to achieve target patient temperature.
Question 4: A perfusionist notices plasma breakthrough (wet gas side) on a hollow-fiber oxygenator 4 hours into a complex case. The BEST immediate action is:
- Increase sweep gas FiO2 to 1.0
- Prepare to change the oxygenator emergently (Correct answer)
- Decrease blood flow rate by 50%
- Add heparin bolus to improve anticoagulation
Correct answer: Prepare to change the oxygenator emergently
Plasma breakthrough compromises gas exchange and risks contamination; the oxygenator must be replaced as soon as safely possible.
Question 5: Which design feature of microporous polypropylene hollow fibers makes them susceptible to plasma leakage over time?
- Excessive wall thickness reducing diffusion
- Open pores that can be wetted by plasma proteins under prolonged exposure (Correct answer)
- Hydrophilic surface coating that attracts water
- Absence of anticoagulant coating on fibers
Correct answer: Open pores that can be wetted by plasma proteins under prolonged exposure
Plasma proteins gradually wet microporous polypropylene walls, allowing liquid plasma to permeate from blood to gas side.
Question 6: When priming a membrane oxygenator, air removal is best confirmed by:
- Visually inspecting the gas outlet for bubbles
- Measuring arterial line PO2 after recirculation (Correct answer)
- Checking that prime volume equals rated priming volume
- Observing clear outflow from venous reservoir drain
Correct answer: Measuring arterial line PO2 after recirculation
Circulating prime and sampling the arterial line for dissolved oxygen confirms adequate de-airing of the fiber bundle.
Question 7: Polymethylpentene (PMP) oxygenators differ from polypropylene oxygenators primarily because PMP fibers are:
- Microporous, allowing direct gas-blood contact
- True membranes with no open pores, preventing plasma leakage (Correct answer)
- Thicker-walled with lower oxygen transfer rates
- Made from silicone rubber for flexibility
Correct answer: True membranes with no open pores, preventing plasma leakage
PMP fibers are diffusion membranes without open pores, eliminating plasma breakthrough and enabling longer safe use.
Which parameter is most commonly used to assess oxygenator gas transfer efficiency during cardiopulmonary bypass?